I have a Hanna 98128, which is giving erratic readings.
I have just now calibrated it (2 point calibration). Then in sample rainwater from a plastic lined tank, it initially read around 6.2, the stability symbol disappeared, but the reading kept dropping for the next few minutes, down to around 5.8.
Rainwater from the other concrete tank initially read 6.8 but it too kept on dropping after the stability symbol disappeared, down to around 6.3.
Dipped the instrument back into the 7.01 solution, and it stabilised pretty well straight away at around 7.03.
If I rinse the electrode, then dip the instrument into the 4.01 solution, it reads pretty well close (4.04), but after rinsing again, the first rainwater sample now reads in the mid 5's. (To rinse, I'm letting running water from the tap I'm about to test from, run over the probe for 20 seconds or so)
So worse than useless (literally) as it stands at the moment.
I'm assuming this is not normal behaviour, so am I not rinsing the electrode well enough between readings, or could the electrode (glass bulb) be on the way out?
I have just now calibrated it (2 point calibration). Then in sample rainwater from a plastic lined tank, it initially read around 6.2, the stability symbol disappeared, but the reading kept dropping for the next few minutes, down to around 5.8.
Rainwater from the other concrete tank initially read 6.8 but it too kept on dropping after the stability symbol disappeared, down to around 6.3.
Dipped the instrument back into the 7.01 solution, and it stabilised pretty well straight away at around 7.03.
If I rinse the electrode, then dip the instrument into the 4.01 solution, it reads pretty well close (4.04), but after rinsing again, the first rainwater sample now reads in the mid 5's. (To rinse, I'm letting running water from the tap I'm about to test from, run over the probe for 20 seconds or so)
So worse than useless (literally) as it stands at the moment.
I'm assuming this is not normal behaviour, so am I not rinsing the electrode well enough between readings, or could the electrode (glass bulb) be on the way out?