That's fine... As you stated your instrument is ATC'd from 0-100°C while most people would have instruments only ATC'd for 10-30°C. I understand that your instrument settles within about 1min, and theoretically the scope instruments should be the same - though maybe this assumption is not correct.Lyrebird_Cycles said:OK I just ran a test on todays forerunnings which are at ~75 oC.
I took a sample and cooled it to 20.0 oC. Three sub samples measured 18.8, 18.7 and 18.8 (all readings in Brix) after approx one minute to equilibrate on the machine.
Three samples taken from the runnings (hot) and each allowed to equilibrate* for approx 1 minute on the machine read 18.8, 18.8 and 18.8.
Equilibration was performed by taking successive readings on the same sample until it stabilised. The machine takes approximately 6 seconds to take a reading then it takes me another second or two to note the reading and press the go button. A typical sequence was 18.0, 18.3, 18.5, 18.6, 18.7, 18.8, 18.8, 18.8, you can see the equilibration actually takes less than a minute.
The stated accuracy of the machine is +/- 0.2 oP but I find it's usually repeatable to within +/- 0.1 on similar samples.
18.8 Brix sounds like a nice beer by the way...
I'll be brewing in the next few days and will do a similar check.