Aces High
Well-Known Member
- Joined
- 30/9/10
- Messages
- 311
- Reaction score
- 50
I thought this would be an interesting experiment. This unit is a metrology calibrator or dry block calibrator. There is a thermowell in the top that you put the thermocouple in and you set the temperature and compare it with the thermo reading
You can set it at any temperature from -45 to 140 degrees with a 0.000 accuracy. its NATA calibrated, so its pretty damn accurate.
These are 2 x STC 1000 units that have been in service for 3 years and about 1.5 years. One controls one of my fermenting fridges, the other controls a HERMS unit
I initially set it for 66 degrees and attached the STC's (sorry, forgot to take a photo) One unit was within 0.1 and the other 0.9 degrees. Not bad from a 30 buck controller. I calibrated them both to exactly 66 degrees using the F4 function which allows up to + or - 10 degrees adjustment. I then set the temp for 17 degrees to check the linearity and I gotta say, its pretty impressive overall
You can set it at any temperature from -45 to 140 degrees with a 0.000 accuracy. its NATA calibrated, so its pretty damn accurate.
These are 2 x STC 1000 units that have been in service for 3 years and about 1.5 years. One controls one of my fermenting fridges, the other controls a HERMS unit
I initially set it for 66 degrees and attached the STC's (sorry, forgot to take a photo) One unit was within 0.1 and the other 0.9 degrees. Not bad from a 30 buck controller. I calibrated them both to exactly 66 degrees using the F4 function which allows up to + or - 10 degrees adjustment. I then set the temp for 17 degrees to check the linearity and I gotta say, its pretty impressive overall